There are some things you need to know about ORP measurement:
1) it's slow, VERY slow and patience is a virtue
2) you need a flow, so at least stir and preferably at a constant speed (and not to slow).
3) CALIBRATE! always do a proper calibration/cleaning before measuring for about every type of electrochemical measurement
4) electrones tend to 'stick' on the measuring part of the electrode (glass membrane with metal bands) so wipe it off before each measurement.
Some usefull information (mostly pH related but there are some ORP hints too):
http://www.consort.be/Support/Techni...ion/APP_pH.pdf
I really recommend reading this document.